Popular Forum Questions
Ask a QuestionThere are no questions. Be the first to ask a question!
Documents
Background and Identification
The Getemed VitaGuard VG 2100 is a heart rate, apnea, and respiration monitor for pediatric and adult patients. Getemed Medezin is a producer of medical devices in the following categories: cardiac function diagnostics, ambulatory patient monitoring, and telemonitoring. The VG 2100 can be used for monitoring vital signs in clinical and home environments.
The VG 2100 monitor features storage and silent alarms for monitored parameters. Additional manual storage of physiological signals increases the number of applications for this model. The monitor features a graphical LCD display that shows various measured parameters, alarm settings, waveforms, and status messages. When physiological parameters move above or below their preset alarm limits, acoustic and visual alarms should occur.
The VG 2100 monitor can be powered by either four 1.5-Volt alkaline batteries or with an external 9-Volt power adapter. With fully charged batteries, the monitor can operate for up to 24 hours.
The name “VitaGuard VG 2100” is printed in blue font in the upper right-hand corner of the device’s front face. The manufacturer’s name, “Getemed'' is printed vertically in white lowercase font on the left-hand side of the device's front face. Four arrow buttons are located on the right-hand side of the device’s display.
Technical Specifications
General:
- Weight with power pack: Approx. 600 g
- Dimensions: 13.5 x 20.5 x 4.5 cm
- Casing material: ABS
- Power supply:
- Power pack (4.8 V, NiMH) or 4 x batteries (1.5 V, LR6, alkaline)
- 9 V power adapter NA 3000-2
- Power adapter NA 3000-2: Input: 100 – 240 V, 50 – 60 Hz, 400 mA / Output: 9 V d.c., 1.5 A
- Power pack recharge time: < 6 h
- Operating time: > 24 hours with alkaline batteries or NiMH power pack
- Replace battery message: Message on LCD display
- Battery exhaustion message: Visual and acoustic warning
- Keys: 6 pushbuttons
- ECG / respiration connector: 8-pin miniature round connector, type CF input
- USB connector: Mini USB to connect with a PC
- AUX connector features:
- RS232 interface for modem or online output
- External alarm unit output
- Nurse call unit output
- 2 analog inputs (AUX 1 & 2)
- Display elements: 5 LED’s and a graphical LCD display (320 x 240 dots) with backlight when powered by external or automobile power adapter
- Alarm warnings: Visual and acoustic as per DIN EN 60601-2-49 and IEC 601-1-8
Apnea Monitor:
- Method: Impedance pneumography
- Respiration range: 4 – 120 breaths /min, depending on age group selected
- Resolution: 1 /min
- Apnea pause settings: 8, 10, 12 ... 30, 32, 34 s
- End of alarm condition: 2 breaths within 6 s
- Minimum amplitude: Approx. 0.2 Ohm
- Signal recognition: Green LED and selectable beep tone
Heart Rate Monitor:
- Heart rate range: 25 – 270 /min
- Bradycardia alarm settings: 30, 35 ... 175, 180 /min
- Tachycardia alarm settings: 100, 105 ... 250, 255 /min
- Minimum amplitude: Approx. 0.2 mV
* Selectable ECG leads: Einthoven I, II or III
Memory:
- Storage functions: Event, trend, full-disclosure and compliance memory
- Storage capacity:
- Approx. 200 events, depending on pre-alarm and post-alarm settings
- 72 h trend, 8 h full-disclosure
- Stored data: ECG, heart rate, respiration, basal impedance, status, 2 analog inputs
Classifications:
- Product classification: IIb as per MDD 93/42/EEC
- Electrical classification: Class II as per DIN EN 60601-1
- EMC classification: Group 1, class B as per DIN EN 60601-1-2
- Ingress protection: IP 21 as per DIN EN 60529
- German “Hilfsmittelnummer”: 21.24.02.4014 (applies to Germany only)
Environmental Conditions:
- Operating temperature range: 5 – 40 °C
- Operating humidity range: 5 – 95 %, non-condensing
- Altitude: -300 m to +5000 m
- Storage temperature range: -40 to +70 °C
