  1. How to Test the Manometer on a Fisher Paykel Neopuff, How to Test the Manometer on a Fisher Paykel Neopuff: step 1, image 1 of 1
    • Lift the cover off the Max Pressure Relief Valve slightly and turn out of the way.
  2. How to Test the Manometer on a Fisher Paykel Neopuff: step 2, image 1 of 1
    • Disconnect all devices from the F&P Neopuff™/Perivent™ outlet port.
  3. How to Test the Manometer on a Fisher Paykel Neopuff: step 3, image 1 of 1
    • Check the manometer needle is within +/- 2 cm H2O of zero on the manometer gauge. If the manometer does not read zero, the resetting of the manometer to zero procedure should be followed.
  4. How to Test the Manometer on a Fisher Paykel Neopuff: step 4, image 1 of 1
    • Connect the outlet of the F&P Neopuff™/Perivent™ Infant Resuscitator to a bleed valve and a certified gauge (e.g. Mensor Digital Pressure Gauge Series 2400).
    • Set the gas supply to 10 L/min.
    • Completely close the maximum pressure limit valve by turning the left-handed knob completely clockwise.
  5. How to Test the Manometer on a Fisher Paykel Neopuff: step 7, image 1 of 1
    • With the bleed valve closed, adjust the Peak Inspiratory Pressure knob to set the pressure so that the certified gauge reads 10, 20, and 40 cm H2O.
    • Check that the manometer reads within +/- 2 cm H2O of these values at each set point.
    • With the pressure set to 40 cm H2O, open and close the bleed valve three times and check the manometer needle rises and falls smoothly.
If the F&P Neopuff™/Perivent™ Infant Resuscitator fails any of these tests, the manometer should be regarded as inaccurate and replaced with a new manometer.