  1. Philips Respironics V200 Esprit Power Supply Replacement, Power Supply: step 1, image 1 of 1
    • Remove the power supply shroud (Instructions found in Chapter 10.8 on page 10-21 of the Service Manual).
  2. Philips Respironics V200 Esprit Power Supply Replacement: step 2, image 1 of 1
    • Loosen the screws on power supply connector TB1 (page 10-22 of Service Manual).
    • Disconnect the following wires:
    • Brown (terminal 1)
    • Blue (terminal 2)
    • Green (Terminal 4)
  3. Philips Respironics V200 Esprit Power Supply Replacement: step 3, image 1 of 1
    • Loosen the screws on power supply connector TB2 (page 10-22 of Service Manual).
    • Red (Terminal 1)
    • Black (Terminal 4)
    • Caution! Do not remove the black ground wire from TB3 Terminal 3.
  4. Philips Respironics V200 Esprit Power Supply Replacement: step 4, image 1 of 1
    • Loosen the screws on the power supply connector TB3 (page 10-22 in Service Manual).
    • Red (Terminal 1)
    • Black (Terminal 2)
    • Caution! Do not remove the black ground wire from TB3 Terminal 2.
    • Using a 4-mm hex driver, remove the eight M5 x 12 screws that secure the power supply to the top enclosure (page 10-22 in Service Manual),
  5. Philips Respironics V200 Esprit Power Supply Replacement: step 6, image 1 of 1
    • For this step, lift the power supply as needed (page 10-22 of Service Manual).
    • Disconnect cables from connectors J1, J2, and J3.
  6. Philips Respironics V200 Esprit Power Supply Replacement: step 7, image 1 of 1
    • Lift the power supply from the top enclosure (page 10-22 of Service Manual).
    • You must install the MMI PCB before installing the power supply. Information on how to do this can be found in Chapter 10.12 on page 10-26 of the Service Manual.
    • When installing the power supply, make sure no cables are routed underneath the power supply.
  7. Philips Respironics V200 Esprit Power Supply Replacement: step 8, image 1 of 1
    • Reinstalling the power supply (page 10-24 of Service Manual):
    • Reinstall the eight screws that hold the power supply to the top enclosure.
    • Do not tighten the screws until all are installed.
    • Connect the following cables to the power supply (page 10-24 of Service Manual):
    • Cable PSU J1 to connector J1
    • Cable PSU J2 to connector J2
    • Cable PSU J3 to connector J3
    • Install the following wires to the power supply (page 10-24 of Service Manual):
    • Red and black wires (PSU TB3) to connector TB3
    • Red and black wires (PSU TB2) to connector TB2
    • Blue, green, and brown wires from the AC mains panel to connector TB1
    • When installing, push the wires completely into the connectors.
    • No wires should be visible after being tightened at the terminal block.
  8. Philips Respironics V200 Esprit Power Supply Replacement: step 11, image 1 of 1
    • Reinstall the power supply shroud (Instructions found in Chapter 10.8 on page 10-21 of the Service Manual).
    • Make sure the red and black wires from TB3 rest inside the shroud slots.
    • Make sure the shroud does nor pinch any cables.
    • All cable assemblies and wires must be routed between the shroud and the top enclosure wall.
    • The black ground wire reattaches to the shroud with the longer screw (near connector TB2).
Finish Line
Grace Gius
505 Reputation
3 Guides authored
Cal Poly, Team S1-G4, Zimmerman Spring 2020 Member of Cal Poly, Team S1-G4, Zimmerman Spring 2020
CPSU-ZIMMERMAN-S20S1G4
5 Members
5 Guides authored
One Comment
Is this available?
Shashikumar G - Reply
Past 24 Hours: 0
Past 7 Days: 3
Past 30 Days: 7
All Time: 364