Hamilton G5 Ventilator Troubleshooting
Troubleshooting guides for the following - Fluid Collecting in Breathing Circuit - Ventilator not Ready for Use - Air Pressure Alarms - Restrictions in the Breathing Circuit - Display Frozen/Not Visible -Flow Sensor Test
Fluid is collecting in the breathing circuit or tube
Fluid pools in the main or ancillary tubes causing the ventilator to not work properly.
Moist air or fluid in the lungs condensates in tubing (usually collects at lowest point in the tube)
Disconnect the tube affected from the front of the ventilator. Use gravity or controlled motion to bring the fluid out of the tube. Once the tube is clear of fluid, reconnect to ventilator and run the “Sensor Flow” test(refer to header below) to make sure everything is working properly.
Note. Disconnecting the tube from the ventilator will not provide ventilation for the patient momentarily. Make sure the patient is taken care of while the tube is disconnected.
Faulty Tubing
Replace tube with a clean one.
Run “Flow Sensor” test.
Ventilator not set up for use
When using a ventilator that is not currently being used, it is important to perform several steps to ensure the ventilator is working properly and ready for use.
Pre-use Tests not up to Date
Start with a “flow sensor” test.
Next, run a tightness test:
- Similar to the “flow sensor” test. Press the system button on the bottom right of the screen, then press on the tab “Tests & Calib”.
- Click “Tightness” to run the test.
- If the ventilator asks to “Tighten the patient system,” place the cap on the ventilator tube end.
- Once the tightness test is done, the display will show “Patient system tight.”
- Uncap the ventilator once the test is complete. There will be a green check with the updated time of the most recent test.
Last, run an O2 cell test:
- Click “O2 Cell” to run the test.
- If the test takes a long time to run, this is a sign that the O2 fuel cell should be replaced soon.
- Once the O2 cell test is done, the display will show “O2 Cell Calibrated OK.” There will be a green check with the updated time of the most recent test.
Air Pressure Alarm
Alarm goes off too frequently and for unnecessary reasons
Improper Alarm Setting Set
On the display, click the “Alarms” button. Adjust top and bottom numbers to set range for various alarms. If the alarm is going off unnecessarily, the lower pressure limit is likely set too high. Consult with your respiratory therapist to confirm a safe limit.
Flow Sensor not Calibrated
Run “flow sensor” test and confirm with a green checkmark.
Breathing Pathway Restricted from “Kinks”
Kinks in tubing can stop the circulation in a ventilator or slow it down. If fluid pooling in the circuit is not the issue after performing a flow sensor test, check for kinks in the tube.
Tube is Bent
There are many pieces of equipment and tubes in a room where people are working and patients are moving around. The ventilator tubes are not rigid and sometimes get bent when they are moved around. Inspect tube for kinks, and remove kinks by squeezing at kink or gently pulling on each side of kink.
Display Frozen/Not visible
The screen is not responding to any buttons being pressed.
Unresponsive Display
Turn the display on and off from the button on the back. If the display still does not work, the display must be replaced - contact Hamilton Medical.
Brightness too Low
Check the ventilator’s brightness by pressing the “System” button in the bottom right of the screen. Then press on the “Day/Night” tab. From there you can adjust brightness settings to make the display more visible. If the display is fading even when brightness is turned up, consider replacing the LCD screen - contact Hamilton Medical.
Running a “Flow Sensor” Test
- Make sure the ventilator is on. If not, press the power button on the back of the console.
- Press the “System” tab.
- Select “Test & Calib” tab.
- Click the “flow sensor” test to run the test.
- Once complete the test should return a green-check and an updated date & time.
- If the test fails, check for kinks or fluid in the tube. If no kinks or fluid are present, consult with your respiratory therapist or technician.
0 Comments